SQM-LU-DL (Datalogging light meter) / Re: 3 new DLs installed...
« Last post by admin on December 10, 2014, 04:02:12 pm »There are two ways to gather readings with the DL meter:
If you noticed that no readings are being recorded at all, then check the threshold setting on the datalogging tab. It should be 0. Anything greater will prevent datalogging on readings below that value.
Also, when plugging in the battery pack to the meter, the LED should be visible for three seconds every minute.
Please let me know if you have further questions or still have problems.
- Log continuous mode: Available from the information tab. This method does not set the internal battery operated datalogging features. If you initially tested the unit with this mode, then the battery operation settings will not be set to the same setting as in Log continuous.
- Datalogging tab: Settings on the datalogging tab are stored in FLASH memory inside the meter. These settings control how the meter records while connected to the battery.
If you noticed that no readings are being recorded at all, then check the threshold setting on the datalogging tab. It should be 0. Anything greater will prevent datalogging on readings below that value.
Also, when plugging in the battery pack to the meter, the LED should be visible for three seconds every minute.
Please let me know if you have further questions or still have problems.